Post by chase on Mar 12, 2023 22:55:32 GMT
A 2008K machine that has not been in use for a while powers up but ends up with an "X" next to each of the messages below.
. Level detector
. Blood leak . Arterial
. Venous
. TMP
. Optical detector
When the RESET button is pressed on the machine, the machine allows normal dialysis to take place.
Question
1. Has anyone experienced this before? 2. What action needs to taken to resolve this so that I get check marks in the box next to each message instead of an "X"?
Thanks.
|
|
|
Post by dave74 on Mar 13, 2023 2:54:57 GMT
|
|
chase
New Member
Posts: 16
|
Post by chase on Mar 13, 2023 4:07:02 GMT
Thank you for your response. However I cannot open the link. When I click on the link I get the message - "Oops, there was an error!"
|
|
|
Post by Chuck Weddle on Mar 13, 2023 11:27:50 GMT
Thank you for your response. However I cannot open the link. When I click on the link I get the message - "Oops, there was an error!" The link just worked for me. Try again, it may have just been a burp.
|
|
|
Post by dave74 on Mar 13, 2023 13:36:29 GMT
I copied and pasted it.
I had a Blue Star machine with about 8k hours that started failing level Det, Blood leak, Arterial, Venous, and TMP. I tried the usual reseating boards and making sure everything was calibrated. Nothing helped, I decided to change a board or 2. Wound up changing sensor, actuator, function, and even power logic just to make sure and nothing helped. By this time I was ready to burn it, but I didn't wanna give in and call tech support just yet, so I changed the mother board just to make it complete, still failing same tests....so, I gave in and called tech support.....and I couldn't believe what he told me, changed out the LP955 board and low and behold, it worked! Apparently if the board doesn't send a signal letting the machine know when the pump starts and stops each time, it will not pass that specific test. I had never ran into this problem before. Hopefully this will help someone out in the future
|
|
|
Post by gnurk on Mar 14, 2023 12:43:58 GMT
when you have this type of problem swap out in this order air detector module, module cable to motherboard this will narrow it down to modules or board in the upper
|
|
|
Post by phillipashleyludlow on Mar 16, 2023 12:44:25 GMT
|
|