Post by SheepDog1 on Mar 14, 2023 17:02:21 GMT
with these new machine i have been getting a lot of SV41 errors. have found that the (L) calibration has been set a little low from set up or has changed with use. simply recal L in service mode. you will need the starter kit from Nikkiso and use the supplied gage. just trying to put up some Nikkiso content since they are fairly new.
|
|
|
Post by nikkisodude on Jun 3, 2023 3:58:48 GMT
you can check your ef-02 filter, filter was contaminated probably.
|
|
|
Post by Sheepdog1 on Jun 8, 2023 17:48:46 GMT
ya replaced that at the beginning and same result. when the machine gets to DM test 29 check the SV41 value (TK Red, monitor, Start up ) scroll down to SV41 test measurement and if its below 100 it will fail and alarm. L calibration is probably low. Re Cal on the higher side and you should be back in Biz
|
|
|
Post by Boobsie on Jul 18, 2023 12:54:03 GMT
Hope you guys live in a place where humidity are low. Where having a lot of issues on Humidity in our area. Just FYI, MB are not cheap to replaced.
|
|
|
Post by sheepdog1 on Dec 13, 2023 1:35:36 GMT
hey bobby what were your symptoms . i got one now, it randomly shut off during treatment and getting word its the main PCB?
|
|
|
Post by sheepdog1 on Dec 13, 2023 1:36:32 GMT
Boobsie
|
|