Post by Unknown24 on Mar 17, 2023 0:48:30 GMT
Hello everyone,
I have a 2008T machine with heat disinfection problem. Machine reaches to 90 degrees and spits out water from the vent tubing. I have replaced the heat exchanger, verified brushes on both pumps. any suggestions?
|
|
|
Post by gnurk on Mar 17, 2023 12:37:43 GMT
go thru ts guide sounds like a board prob function or power logic
|
|
|
Post by jimjamjr on Mar 21, 2023 17:24:19 GMT
Quickest part from TS guide i remember for "Temp over 90" was to re-seat the boards. take them out and put them back in, make sure they have a good connection and see if that works.
as for the water coming out of the back, might need to dive into the TS guide for that to fully be sure. could be high inlet pressure, bad balancing chamber diaphragm, bad float switch, etc.
|
|