Lx 800
Mar 22, 2023 11:46:27 GMT
via mobile
Post by Deleted on Mar 22, 2023 11:46:27 GMT
Good afternoon, Dialog+ , made 2018, soft 9.1b , 8800h , the machine has batteries, they are in good condition. During dialysis the blood pump stops, the alarm sounds, the red LED lights come on, there is no error, the alarm is not reset, the blood pump stands, when the machine is restarted the message "System restored after power failure or disconnection" appears. This situation occurred three times: the first time the error was once per procedure, the second time twice with a 5 minute interval per procedure, the third time 3 times, the error was repeated after resuming the procedure and the patient was moved. What was done:
1)Moved the power supply to another unit ( to kidney number 7) 2) I transferred Lx 800 motherboard with Cf card to another kidney, new battery on the motherboard, it is in good condition (on kidney number 6) The result, on the kidney (number 6) where another motherboard was installed, the alarm began to appear, after rebooting this problem may occur again, again reboot and continue dialysis. My question is: Should I buy a new motherboard? Should I try to replace the Cf card? Reinstall the software?
|
|
|
Lx 800
Mar 23, 2023 13:07:20 GMT
Post by Zombie on Mar 23, 2023 13:07:20 GMT
"system restored after power failure" means you lost AC power to the HD machine. you stated machine has batteries. are you referring to backup battery unit at bottom of machine? verify AC power cord is good. I dont think the power supply or the motherboard is the issue or the CR2032 battery in the motherboard unless your getting a "system error alarm" you can try reloading software
|
|