Post by crarogers on Mar 23, 2023 20:03:55 GMT
How often are you heating a machine while doing 24 hr SLED treatments. Fresenius recommends Q 12 hrs. We do Q 24 hrs, water cultures always negative. Taking a patient off SLED to heat the machine every 12 hrs puts some of them at risk of death or serious complications.swapping out a machine takes time and not always possible due to staffing.
|
|
|
Post by CafeRacer68 on Apr 3, 2023 15:21:37 GMT
when you say 24 hr treatment how are you running past 12 hours? our program locks us out until some type of disinfection is done. we normally switch them out but if we have a few hours to kill while patients getting test/procedure done I will have them bleach so they don't have to wait for the temperature to come down.
|
|
|
Post by CafeRacer68 on Apr 3, 2023 15:22:40 GMT
disregard, just noticed the title said K, I miss those for running sled, the new Ts lock you out after 12 hours
|
|