Post by BioLee on Apr 1, 2023 17:17:42 GMT
Hi, I have a 2008T Bibag Bluestar premium with 3k hours on it. I have an issue when i'm trying to calibrate the temp control sensor. My issue is that in the first part, when the temperature is stable, I can put the temp of the meter on to the machine as supposed. But at the second part of the calibration, when the second timer finishes, I get a type of alarm that says "pre/Post Temp out of range" I have calibrated the pre/post temperature sensors with the resistor a couple of times, and I still got the message when I try to calibrate the machine. Any ideas of what could be wrong? Thanks in advance.
|
|
|
Post by gemo.mad on May 11, 2023 18:34:41 GMT
I had the same problem, I changed the post sensor (M46405). That worked for me.
|
|
|
Post by Unknown24 on Jun 23, 2023 21:35:56 GMT
Any luck?
|
|