Post by BFC on Feb 8, 2022 22:28:58 GMT
Is anyone familiar with the Aquaboss RO system for the acutes program. I have a RO leaking thru the chemical port.
|
|
|
Post by Chuck Weddle on Feb 8, 2022 23:09:06 GMT
Are you talking about the Aquaboss sold by B. Braun or the AquaB sold by Fresenius?
|
|
|
Post by Probably AquaBoss on Feb 9, 2022 12:04:34 GMT
Are you talking about the Aquaboss sold by B. Braun or the AquaB sold by Fresenius? He wrote AquaBoss twice so I think he means the AquaBoss.
|
|
|
Post by Admin on Feb 9, 2022 12:27:25 GMT
Chemical Port?
Can you post a picture please?
|
|
|
Post by BFC on Feb 9, 2022 16:13:54 GMT
Yes I believe it was sold to BBraun, but no longer supported
|
|
|
Post by Admin on Feb 9, 2022 16:17:34 GMT
Can you please post a picture of what you're describing? Are you referring to the portable?
|
|
Deleted
Deleted Member
Posts: 0
|
Post by Deleted on Feb 9, 2022 16:31:18 GMT
|
|
Deleted
Deleted Member
Posts: 0
|
Post by Deleted on Feb 9, 2022 16:32:42 GMT
|
|
Deleted
Deleted Member
Posts: 0
|
Post by Deleted on Feb 9, 2022 16:35:26 GMT
Here you go
|
|
|
Post by Chuck Weddle on Feb 9, 2022 17:16:08 GMT
How do you inject the chemical? Insert a syringe as if were a sample port? If so, I'm guessing it's seal is worn out. You could try flushing it with water, maybe a piece of dirt keeping it from sealing. How old is the machine?
|
|
Deleted
Deleted Member
Posts: 0
|
Post by Deleted on Feb 9, 2022 17:30:34 GMT
Its leaking while running operational mode not durimg disinfect. We opened it up and nothing is loose or disconnected.
|
|
Deleted
Deleted Member
Posts: 0
|
Post by Deleted on Feb 9, 2022 17:38:43 GMT
Nurse mentioned it started squirting out of the chemical port and water coming out of the bottom while in dialysis mode
|
|
|
Post by Anon’s Face on Feb 9, 2022 18:39:07 GMT
Has it recently been disinfected? During the disinfect you typically cap the concentrate hose, has that cap been removed? No errors or faults while running in dialysis mode? Is it also squirting out in standby phase or ONLY in dialysis mode? Is your inlet water regulator adjusted to 18-20 psi on the lower end, ideally peaking to around 50 psi? There’s also a check valve located on the bottom, opposite TS2 that could be failing.
|
|
|
Post by Admin on Feb 9, 2022 18:53:08 GMT
Based on the system - with the water coming out of the block and the chemical port - it can only be the level sensor causing the issue.
|
|
|
Post by Koks on Jun 10, 2022 17:47:12 GMT
Try installing a quick connect with tubing that goes into the tank, the other side you can use for disinfection
|
|