Post by fishnqueen1 on Jun 6, 2022 18:46:47 GMT
Good Afternoon, I have a Bluestar machine that keeps failing all the Actuator Temp. test's High and low, it also fails Actuator Conductivity high and low, only when we try to use the Bibag. When we use the jug it passes everything. iv replaced sensor board, actuator board, both cables . Replaced all temp sensors and calibrated them all, also calibrated conductivity cells. calibrated acid and bicarb pumps, calibrated transducer pressures checked D-Air /Flow / Loading pressures- all good. Has anyone had this problem any input is greatly appreciated. thanks
|
|
|
Post by F a bibag machine on Jun 6, 2022 23:55:38 GMT
I've seen this when running on a wall box but i have yet to figure out the root cause. When i re-run the alarms tests after failing it passes every time. Starting to wonder if its a software issue?
|
|
|
Post by gnurk on Jun 7, 2022 9:50:23 GMT
bibags are a tough nut have you called tech services what do they say
|
|
|
Post by F a bibag machine on Jun 8, 2022 1:28:12 GMT
It's not just bibag machines, have some regular 5 T's that do the same thing at another clinic and didn't start till the 2.74 was installed on them.
|
|
|
Post by Conrad on Oct 20, 2022 0:45:29 GMT
It's that board in the bibag assembly 2 in the middle, bibag distribution board.
|
|
|
Post by jaretac on Oct 20, 2022 20:02:57 GMT
I wonder about Bibag interface board connected to the actuator board.
|
|
|
Post by babybiomed92 on Feb 27, 2023 20:20:12 GMT
Was there any specific fixes to this question? having same issue on my 2008t bluestar machines. Just seeing if anyone has some fixes? doing everything listed in first message. Thank you all!!!
|
|
|
Post by Kou on Mar 22, 2023 16:48:47 GMT
Good Afternoon, I have a Bluestar machine that keeps failing all the Actuator Temp. test's High and low, it also fails Actuator Conductivity high and low, only when we try to use the Bibag. When we use the jug it passes everything. iv replaced sensor board, actuator board, both cables . Replaced all temp sensors and calibrated them all, also calibrated conductivity cells. calibrated acid and bicarb pumps, calibrated transducer pressures checked D-Air /Flow / Loading pressures- all good. Has anyone had this problem any input is greatly appreciated. thanks i am having the same issues too.. on a T-machine.. any ideas?
|
|
|
Post by rt198p on Apr 7, 2023 16:02:40 GMT
Anyone found a fix for the Non-bibag machines
|
|
|
Post by kou on Apr 18, 2023 14:50:48 GMT
Good Afternoon, I have a Bluestar machine that keeps failing all the Actuator Temp. test's High and low, it also fails Actuator Conductivity high and low, only when we try to use the Bibag. When we use the jug it passes everything. iv replaced sensor board, actuator board, both cables . Replaced all temp sensors and calibrated them all, also calibrated conductivity cells. calibrated acid and bicarb pumps, calibrated transducer pressures checked D-Air /Flow / Loading pressures- all good. Has anyone had this problem any input is greatly appreciated. thanks i have the same issue , unable to fix
|
|
BMF
Junior Member
Posts: 85
|
Post by BMF on Apr 19, 2023 16:48:48 GMT
How close to TCD is your actual conductivity? If you are say .3 or so away from TCD, I have seen these temp and conductivity failures.
|
|
|
Post by manof on Apr 19, 2023 17:50:34 GMT
I would suggest reseating actuator board and recall temp comp make sure to turn temp com off after calibration.
|
|