Post by Bubba on Aug 25, 2022 19:53:03 GMT
Bibag machine with 17717 hours. I replaced a UI-Mics board uploaded chairside image, it is working properly. Problem I have now is i go to rinse machine 15 seconds then get check acid connector. Keep in mind this machine had just been through a bleach. I can select chemical rinse and it goes through pre-rinse just fine. select acid rinse goes through pre-rinse with no issues. I put it through a complete heat disinfect no issues. I have checked all cables and reseated the boards. After heat I cut the machine back on now getting V105 stuck.
|
|
stuff
Full Member
Posts: 220
|
Post by stuff on Aug 26, 2022 11:00:53 GMT
Check the wires on the valves. With bibag machines I have had more wires just randomly coming off the valves than years of k/k2 machines
|
|
|
Post by OverEasy on Aug 26, 2022 18:54:52 GMT
Look into transducer 106 readings and make sure its calibrated correctly. Post your fix.
|
|
|
Post by Hopping John on Aug 27, 2022 2:06:24 GMT
It could be a dirty filter in the acid handle.
|
|
Deleted
Deleted Member
Posts: 0
|
Post by Deleted on Aug 29, 2022 14:49:28 GMT
usually if the transducer wont calibrate its 106 or 105 when replacing make sure board dont get wet, if no luck from there you move on to valves, next check cable that goes to the interface board before the board its self
|
|
Deleted
Deleted Member
Posts: 0
|
Post by Deleted on Oct 6, 2022 19:06:24 GMT
its always brushes for me when it says that
|
|
|
Post by gnurk on Oct 7, 2022 12:20:20 GMT
check that your acid connector has the magnet on it and that your reed switch is ok
|
|
|
Post by tron on Nov 16, 2022 15:37:34 GMT
any update on this? i am having the same issue
|
|
|
Post by t1spike on Nov 16, 2022 18:38:31 GMT
first thing to check the wand filter /after clinics switched to Citra clean seems those get plugged faster and will throw that error alarm
|
|
|
Post by F a bibag on Nov 16, 2022 23:18:21 GMT
Try replacing the acid port, I have replaced the acid port when this happens and have not had a single one come back for same issue.
|
|
|
Post by whitechocolate on Nov 18, 2022 19:58:23 GMT
Replace transducer 106.
|
|
|
Post by gfreely on Nov 21, 2022 13:01:24 GMT
Could be your snap ring cannot hold against pressure for long (had this happen recently with the same issue you're having. Replaced ring and worked fine)or 105 may be crapping out.
|
|
|
Post by Tom on Jan 13, 2023 16:33:45 GMT
Try changing actuator test board..
|
|
|
Post by Fishnqueen56 on May 23, 2023 18:06:22 GMT
my fix was the ribbon cable going to the gen2 board that's connected to the Actuator board, it was just about unplugged. I had replaced the keyboard earlier and didn't notice it.
|
|
|
Post by gnurk on May 28, 2023 12:28:12 GMT
thanks for posting your fix
|
|