Post by Guest I on Sept 24, 2022 18:51:58 GMT
I have a 2008T Bluestar with 6500 hrs. I installed a new flow motor calib motor with no problem ran machine for a while pass alarm test many times. Turn machine off push hydraulic carriage back in button it down. When I turned machine back on all hell broke loose, 24V low, blood still sensed, Art. BP no comm. check all connections on all modules cleaned and reset all boards. Seems like I’m overlooking something. Can anyone help out.
|
|
|
Post by gnurk on Sept 25, 2022 12:06:41 GMT
are you getting a e03 error with your blood pump module if so call tech services they will send out a new module free
|
|
|
Post by dave74 on Sept 25, 2022 13:25:33 GMT
It could be a coincidence, but it sounds like something happened when you pushed the hydraulics back in. Did you check all the connections at the distribution board to see if any cables got pulled loose? Is it possible one of the cables got pinched and damaged when you put the hydraulics back in? Did you check the connections leading to the card cage itself?
|
|
jtech
New Member
Posts: 44
|
Post by jtech on Feb 28, 2023 16:33:58 GMT
Having this same issue on a Blue Star machine with about 20k hours. Did anyone figure out what's causing this?
|
|
|
Post by just a BioMed on Mar 2, 2023 18:30:42 GMT
24v Blood sensed Art bp no comm
3 different directions of bad signal. Try the motherboard
|
|
|
Post by fresenius123 on Mar 16, 2023 15:39:32 GMT
Function board.
|
|
|
Post by Leo Villarreal on Oct 19, 2023 19:04:14 GMT
I had a similar issue with Art BP no comm and a code with an E.08. After extensive troubleshooting, found the transformer on the power supply was the issue. Voltage would drop below the tolerance of 22.8v causing the alarm. Replaced the transformer and voltage holding steady at 24-25v. Just something else to look at.
|
|