Post by Deleted on Oct 28, 2022 15:55:11 GMT
Intermittent error they are getting. They told me today it’s been happening for a few weeks. Sometimes they’ll use liquid. Of course they haven’t said a word about it to me until today lol. Any insights as it’s the first time I’ve see this error. The bibag pressure transducer possibly? Wanted to see if anyone had it before I started to stick my big hands down there to change that transducer.
|
|
|
Post by Guest on Oct 28, 2022 17:58:44 GMT
First, try calibrating the regulator transducer in service mode.
|
|
|
Post by MicroTone on Oct 28, 2022 18:49:17 GMT
If you haven't already, you can try a new bibag or replacing the bibag O rings
|
|
|
Post by whitechocolate on Nov 10, 2022 16:03:53 GMT
Attempt to calibrate the regulator pressure first. If it doesn't allow you to do so, replace transducer 110 and then calibrate.
|
|