Hi Guys I have that 2008t machine that have flow error just start dialysis mode, not in rinse or heat or the others, just in dialysis mode. And that happen after upgraded to BlueStar Premium... I replaced all boards, flow and dearation pumps included heads. Water inlet pressure regulator calibrated (and is brand new), acid and bicarb pumps rebuilt and calibrated, heater exchange replaced, valve 29/41 replaced, valve 30 replaced, cables.... all of them replaced. This machine is almost brand new!! But the error is still there. Any suggestion?
|
|
|
Post by gnurk on Nov 3, 2022 12:02:39 GMT
go thru ts guide and insure your loading pressure is ok is it a high flow or a low flow error how many hours if still under warrenty get fmc to fix
|
|
Deleted
Deleted Member
Posts: 0
|
Post by Deleted on Nov 4, 2022 3:22:05 GMT
Check if the orifice is plugged might be why you're only seeing the error in dialysis mode
|
|
|
Post by Maxine on Nov 9, 2022 17:38:34 GMT
on the hardware options screen ..... make sure dialysate sampling is turned off.
|
|