Post by biomedfossil on Nov 14, 2022 15:11:01 GMT
I have a T machine with 30,000 hours I have noticed around the wording (dialysis, rinse and etc) It's yellow for some reason. Any thoughts?
|
|
|
Post by jeff stein on Dec 6, 2022 1:47:39 GMT
reseat the video cable small white cable between Umic board and function board
|
|
|
Post by DY8UGR on Dec 28, 2022 18:28:23 GMT
Check also your UMIC connector pins (that connects to function board), make sure nothing is cut off or bent.
|
|