Post by beavis on Nov 15, 2022 20:23:43 GMT
need some help. 2008t/bibag. i am getting this error and when trying to calibrate the transducers and i am getting an acid error during rinse(rinse shuts off after 15 seconds)
|
|
|
Post by gnurk on Nov 16, 2022 13:02:25 GMT
try to calibrate using jugs, maybe your loop pressure is to high
|
|
|
Post by F a bibag on Nov 16, 2022 23:15:54 GMT
Replace pressure transducer 105.
|
|
|
Post by Cavinator on Jun 20, 2023 16:33:32 GMT
Had this same issue, swapped out Acid Pressure Transducer 106. I also swapped out the board in the assembly due to corrosion. (Gen2 Hydraulics Distribution Board #190860)
|
|