Post by kendec65 on Nov 17, 2022 14:31:59 GMT
I'll keep that in mind and order some new ones. Did they ever have a recall on these or give customers free magnets to exchange so this issue doesn't get this far?
|
|
|
Post by Chuck Weddle on Nov 17, 2022 14:36:55 GMT
I'll keep that in mind and order some new ones. Did they ever have a recall on these or give customers free magnets to exchange so this issue doesn't get this far? I've never heard that they admitted that there is an issue.
|
|
|
Post by kendec65 on Nov 17, 2022 15:54:48 GMT
Well, I ordered new pumps and hope that helps. Tech support had me check 1.13 to make sure the valves work and they all do. Then they had me run 1.9 and it fails step 3. Would this be the pumps?
|
|
|
Post by kendec65 on Nov 28, 2022 11:52:26 GMT
So, I have a question. I had another machine go down over the weekend and one this morning with broken magnets. How do I go about filing a complaint with someone other than Braun?
|
|
|
Post by Chuck on Nov 28, 2022 12:32:05 GMT
Im doing this from the woods while on vacation with a very weak signal. Make sure you save the parts and tell tech service you want to file a PIR. That escalates it to Germany. You can also file a Medwatch report with the FDA.
|
|
|
Post by SrCusEngr on Dec 11, 2022 0:33:20 GMT
Assembly tools 704130 (needed) consists of a pin torx bit (security bit) and a spacer tool for the new magnet alignment. Only need to purchase one time.
Retrofit gear pump kit. 34562310 to repair up to 10 pumps, I believe.
|
|