New Member
Life is Good.
Posts: 25
|
Post by quilla on Nov 21, 2022 20:56:08 GMT
Hello Folks, I have a machine failing at all time the NEG FLOW ON during the negative pressure test. That is the ONLY test failing. Any thoughts , I will appreciated,
Thanks,
|
|
|
Post by gnurk on Nov 22, 2022 13:28:29 GMT
check your pressures, do a induced positive pressure test let run for 10 minutes check for leaks, do a negative pressure test ramp up the tmp to over 400 use flashlight to look for air bubbles, check your dialysate pressure calibration
|
|
stuff
Full Member
1111111111111111111111111111111111111111111111111111111111111111111111111111111111111111111111111111
Posts: 228
Member is Online
|
Post by stuff on Nov 22, 2022 13:54:30 GMT
I agree with gnurk. Step 1 is to check pressure on both degas and flow motor. This could be as simple as brushes changed out or a frustrating hunt for bubbles
|
|
|