Post by whitechocolate on Nov 28, 2022 16:09:40 GMT
2008T Bluestar with 11,xxx hr
Bibag is overfilling with no air gap. Tech services suggested v100, ribbon cable, or interface board...
So far I have swapped bibag assembly, bibag interface board, swapped actuator test, function, and interface boards from another good machine. the good machine has air gap in the bag..
Machine runs correct conductivity, correct bag pressure, but overfills the bag..
Any other suggestions?? The only bibag parts I haven't swapped is the door assembly and the 104/105 assembly..
|
|
|
Post by MicroTone on Nov 28, 2022 21:12:05 GMT
Are you getting an error banner?
|
|
|
Post by gnurk on Nov 29, 2022 13:07:02 GMT
have you called tech services
|
|
pacnw
Full Member
Posts: 177
|
Post by pacnw on Nov 29, 2022 14:31:54 GMT
Only time I’ve seen this the #1 and #2 hoses going to the bibag door were reversed. Don’t trust the labeling on the hoses as I’ve also found them to be incorrect from the factory. Follow them to verify
|
|
|
Post by whitechocolate on Nov 29, 2022 16:07:08 GMT
Only time I’ve seen this the #1 and #2 hoses going to the bibag door were reversed. Don’t trust the labeling on the hoses as I’ve also found them to be incorrect from the factory. Follow them to verify Nailed it! That was the issue. Thanks for the suggestion!
|
|
pacnw
Full Member
Posts: 177
|
Post by pacnw on Nov 29, 2022 18:45:23 GMT
Happy that worked for you!
|
|
|
Post by gnurk on Nov 30, 2022 14:32:43 GMT
thanks for posting your fix
|
|