v105
Dec 2, 2022 12:00:53 GMT
Post by biobill on Dec 2, 2022 12:00:53 GMT
I am new to the bibag machines, when I change from the wall to jug acid, I get a v105 stuck closed , Is this normal, and how do I fix it
|
|
|
v105
Dec 2, 2022 13:18:16 GMT
Post by gnurk on Dec 2, 2022 13:18:16 GMT
put machine in rinse it should clear, when switching from wa;ll to jug turn flow off first
|
|
|
v105
Dec 7, 2022 13:35:43 GMT
Post by whitechocolate on Dec 7, 2022 13:35:43 GMT
First make sure that the wand you are using isn't the issue. I've seen that happen before. Does the machine run correct conductivity even with the error? If it does, transducer 106 could be the issue. If it does not, the machine isn't drawing acid from the jug.
|
|
|
v105
Dec 13, 2022 18:12:05 GMT
Post by Rocket on Dec 13, 2022 18:12:05 GMT
We sometimes run into this at the Hospitals on the jugs. Make sure the wands are not blocked and the wands are not too tight on the jug, it needs to vent as the HD machine removes the fluid from the jug. Your loop is slightly pressurized, and obviously a jug is not. Also make sure the pressure transducers are self-calibrated (only takes a few seconds).
Good luck!
|
|
|
v105
Feb 4, 2023 14:08:26 GMT
Post by briancanufixthis on Feb 4, 2023 14:08:26 GMT
Transducer 106 is just so much fun to swap out. I hope never to meet the guy who designed the sub panel enclosure that involves this repair!!
|
|