Posts: 223
|
Post by joey on Dec 20, 2022 14:53:23 GMT
Hey I haven't had much time to diagnose because we are opening a new hospital but I have a version 8 that wont let water through the machine. it gets to the tank and the float is working right but still no water through machine. I tried pulling up acid in 1.15 and that didn't clear any air out at all. I changed the gear pumps because they were all 16 years old and that did not fix it . Any help would be appreciated!
|
|
|
Post by Admin on Dec 20, 2022 15:51:36 GMT
Hey Joey! Getting water out to fbk1?
|
|
|
Post by kolev77 on Dec 21, 2022 5:44:23 GMT
Hi, check a pressure reduce valve for debris and stuck. Also check dmv valve, down stream to the upline tank.
|
|
joey
Full Member
Posts: 223
|
Post by joey on Dec 21, 2022 16:24:50 GMT
I have water going to tank ...all that is working but it does not go up to the next level so I have nothing at fbk1
|
|
joey
Full Member
Posts: 223
|
Post by joey on Dec 21, 2022 18:20:15 GMT
Never mind I changed a temp sensor and even though I checked it 72 times to make sure the oring was in there before I screwed it back in, it escaped before I actually did so.
|
|
|
Post by Admin on Dec 21, 2022 19:18:10 GMT
Ahhh that will do it! Glad you got it fixed man. Great job!
|
|