I have a version 8 machine that has a problem during disinfection. Solution intake failed and I noticed that temp sensors are not in the same range.
PE reading during disinfection was -744 mmHg and EP 1500rpm.
PE reading during treatment was -746 mmHg and EP 1891 rpm.
|
|
|
Post by Myth21 on Dec 30, 2022 3:56:30 GMT
I would check the RVE valve for the blockage
|
|