Post by quilla on Jan 25, 2023 20:38:24 GMT
Hello Folks, I have a machines that is not passing the Bloodside test during preparation, Any thought?
Thanks
|
|
|
Post by Admin on Jan 25, 2023 21:16:56 GMT
What part of the blood side test? Software version?
|
|
quilla
New Member
Life is Good.
Posts: 25
|
Post by quilla on Jan 26, 2023 21:05:45 GMT
it is software version 9.1 and the machine clamped the venous line, but it cant created the >500 mmHG pressure at PV.
|
|
|
Post by Admin on Jan 26, 2023 21:24:49 GMT
When it clamps, put a tubing clamp under the clamp. Does the pressure go up and pass?
|
|
|
Post by Chuck Weddle on Jan 26, 2023 21:47:37 GMT
Causes we have seen are: 1. Blood pump roller. 2. Leaking venous luer connector. 3. Leaking A-V level block or A-V Pressure Monitor whichever you have.
Does the venous POD fill with saline?
|
|
|
Post by Admin on Jan 27, 2023 1:20:59 GMT
Blood pump roller is a good one, or SAKV isn’t set to the right gap.
|
|
|
Post by bcjammerx on Jan 28, 2023 17:05:27 GMT
just out of curiosity (and since everyone beat me to the good answers XD ) how old are the blood lines you're using? sometimes I put a set in my desk drawer for re-use and after like a week they'll fail.
|
|
|
Post by yourmrfixitva on Jan 30, 2023 15:38:21 GMT
I have a clinic that overtightens the venous line to the PV manometer causing it to crack. How do I know they overtighten them? I have an insider !! It either causes bloodside to fail or tmp issues. It normally cracks at the nut on the inside and as soon as you try to remove the nut it will snap in half.
|
|
quilla
New Member
Life is Good.
Posts: 25
|
Post by quilla on Jan 30, 2023 18:25:22 GMT
Folks , thanks for all the replies. The Venous manometer was leaking . I replaced, and the problem went away.
|
|