Post by Confused on Feb 21, 2023 14:40:56 GMT
So I have a 2008t machine with cons offset failure. Followed troubleshooting, checked 12v reads 11.9. Changed sensor, actuator, function board and eeprom. Still can’t do cond cell calibration. Wires look good, nothing looks out of place. I am lost. The machine was originally pulled, without a tag or anything, for a leak and a Pct unknowingly put it into heat. When I found the machine, it had been spraying hot water inside for at least 20+ minutes. I fixed the leak, turned the machine on and got the cond offset failure at 50%. Since then, nothing has worked.
|
|
|
Post by confused on Mar 31, 2023 19:42:48 GMT
i found the issue, it ended up being a damaged actuator test board
|
|
|
Post by dave74 on Mar 31, 2023 20:39:57 GMT
Thank you for posting your fix.
|
|