Frist time coming acrossed this issue. 2008T machine 20,000+ hours first treatment was fine, started in the middle of second patient of the day. Staff change bibag, but still getting alarm, no visual leaks. Any pointers?
|
|
|
Post by vaiomed on Mar 12, 2023 19:16:14 GMT
I would first calibrate your pressure regulators, if you get a bicarb pressure calibration error, replace your pressure transducer #110. If it passed and still having an issue, check your bibag distribution board for corrosion and reseat your cables going to and from your interface board.
|
|