Post by kts on Mar 17, 2023 18:09:26 GMT
Good day all,
I'm Having trouble with a 9x machine it was giving me a start blood pump message, I pushed the BP start button and nothing. I thought to myself self this is either a BP issue or a Front panel board issue lets try the easier of the 2 first, so I changed the BP and now I'm getting blood pump button stuck alarm. So I changed the FPB to no avail I get the same damn Message the bp starts to rotate then I get the BP button stuck message and the BP stops.
Anything else that I need to look at or am I just blind and can't see what's right in front of me?
Thanks in advance.
Oh I also forgot to mention I had just replaced the Power supply because it wasn't working and I was doing a machine checkout when I came upon this problem.
|
|
|
Post by Guest68 on Mar 17, 2023 18:27:36 GMT
Make sure PX7 and PX8 connectors to the power supply are pulled into the correct port on power supply because the connectors are identical and will plug either way.
|
|
|
Post by Travis CHBT on Mar 20, 2023 19:19:46 GMT
Something that could be overlooked is the buttons being installed incorrectly on the FPB. I've installed them incorrectly before! You can use the sound/button test in TSM to make sure everything is working correctly before going into therapy.
|
|
|
Post by kts on Mar 27, 2023 15:17:58 GMT
Thanks all sorry for the late response, it turned out that it was the FPB I had an abundance of bad boards that were stashed away and no one had marked them
|
|
|
Post by SrCusEngr on Apr 2, 2023 17:21:05 GMT
Sounds like you need to set aside one machine for testing just to find out if your spares are good or not!!!