Post by rt198p on Mar 20, 2023 14:52:26 GMT
Have a machine with a red banner up top saying conductivity low. Conductivity on machine is reading 13.4 and TCD is 13.5 TCD limits are 13.0-14.0. Conductivity is within range but still showing the red banner up top. What could be causing this?
|
|
eric
Full Member
Posts: 111
|
Post by eric on Mar 20, 2023 15:34:42 GMT
is this a T machine? if so, it takes the machine a minute after getting into conductivity to ensure that it is stabilized before it will allow the machine to run without being in bypass. This is a function of the software.
|
|
|
Post by rt198p on Mar 20, 2023 15:48:47 GMT
Yeah its a 2008T non bibag its been on with the banner for over 30mins that cant be normal this is the only machine that has done it
|
|
|
Post by Sheepdog1 on Mar 20, 2023 20:16:23 GMT
re-set your boards
|
|