Post by Jaz on Mar 22, 2023 16:36:08 GMT
I have a version 8 machine. The machine every now and then will go into low temp. It will run fine for a week or even a month and then suddenly low temp. It was running in the low 20s and i put new fuses in and it ran fine for a month, the machine was pulled this week because of low temps in the 20-30s. I ran it in the back and all the temps are normal now. What would cause the temp to go in and out, but not all the time?!
|
|
|
Post by qman213 on Mar 22, 2023 16:46:19 GMT
Have you tried calibrating all temp sensors? Maybe one is right on the edge or something. Without knowing which sensor, it's kind of a crapshoot.
|
|
|
Post by Admin on Mar 22, 2023 17:07:32 GMT
Degassing pressure could be a cause of this.
|
|
|
Post by Jaz on Mar 28, 2023 17:30:42 GMT
Degassing pressure could be a cause of this. For the degassing pressure....Do you think it could be my FPA or FPE pump?
|
|
|
Post by Admin on Mar 28, 2023 18:08:34 GMT
Nope, I'm thinking either EP is weak, or a faulty RVE.
|
|