Post by kts on Mar 30, 2023 12:12:54 GMT
Good morning all,
I have a version 8 machine that temp stays at 27 or so. I recalibrated the temp sensors all calibrated fine when I go to do a final checkout of the machine the heater doesn't turn on at all. Any ideas where I should look?
|
|
|
Post by Chuck Weddle on Mar 30, 2023 12:48:23 GMT
My guess is EP or RVE. The heater won't turn on with inadequate degassing pressure.
|
|
|
Post by Admin on Mar 30, 2023 12:54:10 GMT
Does the heater turn on in TSM? If it does - you have something up with your degassing pressure like Chuck said
If it doesn't turn on in TSM - you have either a bad heater or a fuse is blown
|
|
|
Post by kts on Mar 30, 2023 13:15:31 GMT
Thanks guys that's exactly what it was the pressure sensor
|
|