To all my fellow FTM's... Not 100% sure how many are left to start EMR, however one thing to note... If your machines WILL NOT transfer over BP's from Machine to Renesan, or in my case transferring over incorrect BP readings, one thing you should look at is your IT setup in the back. On the server, the top is for the Ethernet hookup to the wall, the bottom being the actual PT station its suppose to reflect. So if the ethernet number is 1A on the floor, but its PT station 9, you would go 1 up top, 9 on the bottom. Had a field day (really week) with IT trying to figure it out. Wanted to just let everyone know my findings in the event you may be struggling.
|
|