Post by GoBlueBaby on Aug 10, 2021 12:26:51 GMT
Had no temp and it was stuck at 29.2 which is what my tap water is. FMC tech came out changed out power logic and control board, still no temp but was giving power to heater rod. Changed out CFS transducer and now he calibrated temp control of which was unable to do so before the change. Passed all tests RTS
|
|
|
Post by dave74 on Aug 10, 2021 13:31:56 GMT
So was the original problem a flow error, or were all those parts actually bad?
|
|
|
Post by GoBlueBaby on Aug 16, 2021 9:29:02 GMT
So was the original problem a flow error, or were all those parts actually bad? All those parts was bad changed out 1 at a time except power logic and control boards they must be changed out together or they will eventually go bad if not done so. There never was a flow error.
|
|
|
Post by dave74 on Aug 16, 2021 13:04:44 GMT
Thank you.
|
|