Post by LG on Aug 11, 2021 21:08:20 GMT
Checked my TMP,and it was off a little.When I attempt to calibrate dialysate pressure,I can cal at zero,but -250,always get operator error.Have changed #9 transducer already.Dont recal having this issue before..until after 2.72 upgrade.Any ideas,please?
|
|
|
Post by gnurk on Aug 12, 2021 12:28:28 GMT
insure you are following the instructions I thought that you needed to turn the flow off in dialyze before you enter service mode I cant remember
|
|
|
Post by Guest on Aug 12, 2021 21:21:13 GMT
LG
I'm having the same problem with the new 2008T Bluestar. For awhile there I thought i was loosing my mind. Spent half a day tying new things I even change the The transducer #9 three times and same BS. Tomorrow I will try a new Sensor board..
|
|
|
Post by whitechocolate on Aug 13, 2021 12:43:55 GMT
I have Bluestar machines and have come across this issue a few times. Sometimes just turning the machine off and back on will fix it. If that doesn't work, try reseating all boards in card cage. If that doesn't work, replace sensor board. If sensor board doesn't fix it, replace function board.
I've found that the only time I've had to replace the transducer #9 is when my Bluestars fail tmp and optical detect in self test. Sounds crazy but that's been the fix every time.
Also make sure that the transducers you are using are not the recalled versions! 370506-22 is the correct part number.
|
|
|
Post by gnurk on Aug 13, 2021 13:56:20 GMT
has anyone thought to call tech services as these parts could be under warrenty
|
|
|
Post by DIATECH on Aug 19, 2021 21:35:33 GMT
Placed HD System into Dialysis Mode with 500 ml/in QD. Then, calibrate your Dialysate pressure. You are getting that error because your QD is not in 500ml/min during calibration.
|
|
|
Post by lisagalvan on Aug 20, 2021 0:40:16 GMT
I would have the flow at 500,but it would not let me calibrate at -250.I called tech support and he said try to alternate the flow from 500 to 800 and then back to 500.Did not work.Jyst have to keep going in and out of the calibration mode until the value in the box on the calibration screen is below 120 or so.If that value is low enough,I have no problem completing the calibration.Thanks for all the imput!Valuable information!
|
|
|
Post by gnurk on Aug 20, 2021 12:59:31 GMT
thanks for posting your fix
|
|