Post by art1 on Jun 2, 2023 21:31:24 GMT
Hello everybody any suggestion how to fix sound led test alarm fail. I replaced new speaker and also brand new FPB. Sound Led test is still failing. Please help!!!
|
|
|
Post by art1 on Jun 6, 2023 16:41:42 GMT
When I press the wrench icon and the loudspeaker is LS Pulse S -off. LS current S-off. how do i turn on the LS current to On?
|
|
|
Post by Zombie on Jun 6, 2023 18:06:24 GMT
here is repair instructions please refer to LLS test code 16001-16028 after failure to pinpoint error
5.3.5 Alarm System Test Code: 16 Repeats until Error Message: 1 ID Error Message: 1167: Sound + LED test failed Actions after Error: None (test is repeated immediately) LLC Test Code LLC Test Steps LLS Test Steps LLS Test Code 1. – Prerequisites for Test: No alarm activation – – 2. – Test code = 16 to LLS Test code = 16 back to LLC – 3. 1610x OSD LED OFF, Sound OFF No alarm activation Start condition for the test: 1. “OSD-LED OFF” detected 2. “Sound OFF” detected 16001 Timeout 60 s Activation of the red OSD LED 16002 – Wait for feedback of the activation and switch-off of the LED 16003 – Wait for LED off 16004 4. – – Acknowledge with switch-off of the power supply trigger, i.e. switch-off of VGD and VGB 16005 – 16010 5. 1620x Wait for D24OFF and B24OFF (timeout 60 s since start of code 100) – – 6. – – Trigger ON 16011 7. 1630x Wait for D24ON and B24ON (timeout 10 s) – – 8. 1630x D24ON and B24ON: Alarm activation LLC/TLC Activation test alarm with sound and OSD LED red (ID 1002) Condition for the test: 1. “OSD LED red left and right ON” detected 2. “Sound ON” detected 16021 9. – – Acknowledge with switch-off VGD 16022 – 16027 10. 1640x Wait for D24OFF (timeout 60 s since start of code 200) – – 11. – – VGD ON 16028 12. 1650x Wait for D24ON (timeout 10 s) – – 13. 1699 Completed – –
|
|
|
Post by M on Jun 6, 2023 19:52:47 GMT
Try replacing the ribbon cable between power supply and digital board, I think it’s a 12 pin. Common problem if the speaker and fpb are not faulty If that doesn’t sort it out maybe digital board?
|
|
|
Post by SrCusEngr on Jun 11, 2023 3:32:33 GMT
Note, Wrench screen page two, upper left of screen you will see VD24 and VB24. Values should be ON. During test 16000, you should see these switching. Learn when they should switch and monitor. You may have a bad relay in your SMPS.
|
|
|
Post by art1 on Jun 14, 2023 19:08:43 GMT
During test it doesn't switch at all. How can I check if i have a bad relay??
|
|
|
Post by SrCusEngr on Jun 15, 2023 22:02:55 GMT
Pull SMPS out and ohm them. Surprise, the relays, 1F capacitors, sonic alarm and removable fuses (not soldered) are all repairable items on the SMPS. But <M> might be right, it may be the digital board.
|
|
joey
Full Member
Posts: 227
|
Post by joey on Jun 16, 2023 14:02:50 GMT
Pull SMPS out and ohm them. Surprise, the relays, 1F capacitors, sonic alarm and removable fuses (not soldered) are all repairable items on the SMPS. But <M> might be right, it may be the digital board. One time it was power board number 1 ....thanks to SrCus on that one!
|
|