Post by Rhb on Aug 15, 2021 23:22:57 GMT
Havent seen in errors but vent tube Continues to spit. Heat exchanger maybe float switch? Any other suggestions?
|
|
|
Post by anon’s face on Aug 16, 2021 11:18:12 GMT
Havent seen in errors but vent tube Continues to spit. Heat exchanger maybe float switch? Any other suggestions? 99% of the time if it’s happening in heat only it’s a hole in your heat exchanger. Won’t show any errors as well.
|
|
|
Post by gnurk on Aug 16, 2021 12:23:27 GMT
a small amount of water is also normal
|
|
|
Post by dave74 on Aug 16, 2021 13:07:35 GMT
The other part to check is the inlet pressure regulator. If the pressure spikes to the pressure in your water loop when the machine is in heat disinfect, water can leak out the vent hose.
|
|
Deleted
Deleted Member
Posts: 0
|
Post by Deleted on Aug 19, 2021 19:57:54 GMT
Just had this happen to me the other day at one of my hospitals. New heat exchanger and the problem was fixed. Hope it helps!
|
|
|
Post by DIATECH on Aug 19, 2021 22:08:57 GMT
You have a hole in your heat exchanger. Replace heat exchanger and water inlet regulator then call it a day.
|
|