New Member
B Braun
Posts: 41
|
Post by Adnan on Aug 20, 2023 6:08:33 GMT
Hello, anyone encountered this issue before? I have checked the coupling O-rings and the dialyzer and they were fine.
|
|
|
Post by Admin on Aug 20, 2023 12:04:13 GMT
Check the conductivity sensors for leakage
|
|
|
Post by M on Aug 21, 2023 14:21:01 GMT
Also If your unit only uses liquid bicarb and never bicarb cartridge it is possible that a build up of bicarb in the bicarb holder is causing air to be drawn in at the bicarb holder seals which are no longer sealing correctly. Or even a build up of bicarb on the t piece just before the bicarb pump can also cause air ingress Small cracks on the condo sensors at the o-rings are usually the cause though
|
|