Post by texasbiomed on Sept 4, 2023 15:53:50 GMT
T Machine with a Conductivity Low error while conductivity is within range. The bottom parameter line blinks off and on during this error. Has anyone seen this before?
|
|
eric
Full Member
Posts: 111
|
Post by eric on Sept 5, 2023 17:03:49 GMT
At the beginning of treatment there is a delay in the software that waits for the conductivity to stabilize. Is this during treatment or at the beginning of the machine coming up in cond?
|
|
|
Post by texasbiomed on Sept 6, 2023 13:48:17 GMT
eric, this is during treatment. Machine will run fine for a while then all of a sudden, alarm goes off even though conductivity is within range. What I did notice during that time, pre-conductivity cell on debug screen 0 reads 0 when this is happening. Making me think that it's a possible bad cell or acid pump.
|
|
|
Post by dave74 on Sept 6, 2023 15:27:57 GMT
In looking at the picture, I don't see a yellow bar for the lower alarm limit. Am I missing something? Shouldn't there be a yellow bar at 13.1?
|
|
eric
Full Member
Posts: 111
|
Post by eric on Sept 8, 2023 17:54:24 GMT
In looking at the picture, I don't see a yellow bar for the lower alarm limit. Am I missing something? Shouldn't there be a yellow bar at 13.1? Looks like he said its blinking on and off...... just didnt catch the picture when it was on......? Can you get it to happen when you are testing or just when patients are on? Sounds like you are on the right track with the cond cell or a frayed wire......
|
|
|
Post by texasbiomed on Sept 11, 2023 16:14:16 GMT
Replaced both conductivity cells, a known working acid pump, and conducted a balancing chamber leak test and it all passed. Tested twice and it passed all tests. Was tested again today for good measure and we're back to square one. Still has the same alarm and conductivity drops to 10 now and stays there.
|
|
eric
Full Member
Posts: 111
|
Post by eric on Sept 11, 2023 17:15:02 GMT
Is this on a jug or loop?
Is there still cond. on debug screen 0 or is it 0 on debug........?
|
|
|
Post by dave74 on Sept 11, 2023 20:58:46 GMT
Have you taken a sample from the drain hose to see if the conductivity is actually low or if it's a monitor problem? What do the pre and pos cells read on the debug screen? Do they agree?
|
|
|
Post by texasbiomed on Sept 12, 2023 15:27:10 GMT
Pre-conductivity is normal for a while then drops to 10.0. Post-conductivity cell reads normal. Normal as in 13.6-13.7. Double checked diaphragms on balancing chamber and no rips found. Installed new ones anyways for good measure. Will be swapping boards when I get back to the facility.
|
|
eric
Full Member
Posts: 111
|
Post by eric on Sept 13, 2023 14:49:35 GMT
Could have a 'bad out of box' pre cond sensor........... I would try one more new on the pre just to make sure......... otherwise i wonder if the distribution board has a bad connection........ Sorry for the frustration.......
|
|
|
Post by Boobie Miles on Jan 29, 2024 22:39:20 GMT
Always check deaeration/loading pressures before jumping to the next step with conductivity issues. That solves the issues there
|
|
|
Post by gnurk on Jan 30, 2024 13:11:37 GMT
have you called techservices if you have a bad cond cell out of the box they can send you another for free
|
|