Post by tdana3 on Sept 11, 2023 18:30:34 GMT
I have 20x 2008T serial number series 6T0S with Bi-bags and machine hours ranging from 28,000 to 30,000 hours. I completed the BlueStar upgrade on them all with no problems, except for one machine... it won't let me complete the Temperature Control Calibration-Option 2, because unable to get stable temperature.
I've called tech support and they've had me replace, (I don't remember the specific order) over the last 2 to 3 weeks, I've replaced: NTC-1, NTC-2, NTC-3, Pump Heads on Deaeration and Flow motors, Sensor Board, Power Logic Board, Function Board. After each part was replaced, I'd try to calibrate the Temperature Control Calibration-Option 2, with no change.
Anyone have any suggestions please?
|
|
pacnw
Full Member
Posts: 187
|
Post by pacnw on Sept 12, 2023 14:38:42 GMT
are you calibrating this while up in conductivity? are you using jugs or plugged into the wall port? I'd try it on jugs first and then try calibrating it with the acid and bicarb wands plugged into the front of the machine....
|
|
Icebonez
Junior Member
What is better - to be born good, or to overcome your evil nature through great effort?
Posts: 55
|
Post by Icebonez on Sept 12, 2023 15:45:46 GMT
You mentioned that you replaced the Flow Motor pump head, but did you replace your flow motor brushes? In my experience, and from what I have heard from other biomeds, unstable temperature issues can point to an issue with the flow in the machine.
Additionally, I would check your heater bar and see if it ohms out correctly. Sometimes the simplest solution is the correct one.
|
|
|
Post by briancanufixthis on Sept 22, 2023 13:48:45 GMT
Find the cause? Was just wondering how you made out...
|
|
|
Post by Ihadthishappen on Feb 26, 2024 20:35:09 GMT
When this happened to me, I had a bad bc diaphram. It didnt show up until I needed to do the 100 flow rate. My guess is the added time of volume in the chamber caused the tear to open just enough.
|
|