dcvd
New Member
Posts: 18
|
Post by dcvd on Oct 10, 2023 13:15:39 GMT
Do you all use the ESA612 that Baxter sells for the electrical safety. When I took over program I honestly do not think the prior Biomed was doing ES testing. He had a SA2001- Safety Analyzer from BC Biomedical at the clinic. Can you do all the required manufacture electrical safety test with it. If my understanding is correct, you cannot do the PET with this one analyzer.
|
|
|
Post by biomedpete on Oct 11, 2023 16:36:11 GMT
I am guessing you are talking about the Patient Leakage test. Not certain of the capabilities of the SA2001, but if you are able to complete a Chassis leakage with it you can complete the Patient leakage test. Assuming you have the coupler for the Hansen's with the test point terminal. Point of thought, if your uncertain how to complete test you shouldn't really say the other person wasn't doing it.
|
|
dcvd
New Member
Posts: 18
|
Post by dcvd on Nov 16, 2023 17:26:22 GMT
I am sure they were not doing it that is why I said it. There is no Patient Leakage Test Fixture here part # 515037000 when I took over.
|
|