Post by Speedbump76 on Nov 1, 2023 18:45:31 GMT
I have a machine that I replaced V105. Afterwards the machine started with the flow going through the blue port not the blue dialysate line. Both motors are turning the correct direction. I did replace valves 24 and 25 thinking that one of them was sticking open. Machine is not giving any flow errors. It will rinse and bleach but it will not pass the diasafe test. Checked all connections to see if something was reversed and everthing is identical to another machine. Could this be a board issue?
|
|
|
Post by dave74 on Nov 2, 2023 5:22:55 GMT
It seems weird that the problem developed after you replaced valve 105. Are you sure the valve you installed is working properly?
|
|
|
Post by gnurk on Nov 2, 2023 12:07:58 GMT
check also v104 and v108 as you may have disconnected a valve on replacement also check your distribution brd to insure everything hooked up correctly
|
|
|
Post by Speedbump76 on Jan 25, 2024 20:00:02 GMT
Sorry for the late post but I pulled a rookie move and had the tubing switched.
|
|
joey
Full Member
Posts: 227
|
Post by joey on Jan 25, 2024 20:25:59 GMT
That's cool if you run it backwards the hour clock will go down .....I learned that from watching Ferris Bueller's day off
|
|
eric
Full Member
Posts: 111
|
Post by eric on Jan 25, 2024 21:07:22 GMT
That's cool if you run it backwards the hour clock will go down .....I learned that from watching Ferris Bueller's day off Nice, its like the Dialysis Time Machine!!!!!! Thanks for posting your fix! Eric
|
|