Post by Corewellhealth on Nov 10, 2023 16:47:37 GMT
I have heard that all machines in the acute setting must disinfect the machine intake line at least monthly. Is this true and where can I find the standard to show my boss?
|
|
|
Post by Chuck Weddle on Nov 10, 2023 16:58:26 GMT
Refer them to V-220 and AAMI RD52:2004 7.1
|
|
|
Post by dave74 on Nov 10, 2023 17:47:32 GMT
In Annex G.4, Special considerations for acute hemodialysis, AAMI also suggests using opaque tubing if it is approved by the manufacturer.
|
|
cane
New Member
Posts: 27
|
Post by cane on Nov 17, 2023 16:03:15 GMT
This is true, and if you contact Medical Solutions in Kansas City their owner Jack Dillon is working on a machine make this process easy. With many portables being heat disinfect now meeting this standard can be difficult, but the machine at Medical Solutions simplifies it quite a bit.
|
|
|
Post by SrCusEngr on Nov 19, 2023 2:59:40 GMT
So CIP devices are antiques now? Just asking.
|
|
|
Post by dave74 on Nov 19, 2023 21:42:27 GMT
I have deleted a post in which I quoted from the "Acute Dialysis Survey Readiness Handbook". There are a couple of errors in this handbook. The handbook is included with ISO 23500, but it appears it was not formally reviewed by the AAMI Renal Disease and Detoxification Committee.
|
|
|
Post by Chuck Weddle on Nov 20, 2023 0:24:16 GMT
The lead author of that handbook has been known to try to pass her opinions off as regulation.
|
|