Keep getting this alarm. We have a 12 station pod with central delivery system for bicarb and acid. It happens on 3 or 4 stations. Any suggestions?
|
|
|
Post by Chuck Weddle on Jan 5, 2024 17:31:29 GMT
First thing is to make sure the ratio is set correctly in TSM. Next, try a temperature and conductivity calibration.
|
|
btech
New Member
Posts: 43
|
Post by btech on Feb 26, 2024 7:14:39 GMT
After you have checked the things Chuck adviced you too. Check the performance of your FPE pump. If the FPE is not able to supply enough water in to the DF block for two BC cycles you will get the ratio alarms aswell.
|
|
joey
Full Member
Posts: 227
|
Post by joey on Feb 27, 2024 14:52:03 GMT
I would also check the connections to the central supply. Sometimes the concentrate nipple gets groves cut into it over time which lets air through. First thing I would do is look at those stations and see if air is moving through the lines during treatment. I would also put those machines on jugs and see if it still happens. Also I would check/change the orings in the couplers. Keep us posted.
|
|