Post by spartaninohio on Jan 9, 2024 15:24:34 GMT
Machine failing venous alarm tests, calibrated venous and arterial pressures and tried putting in new actuator board and inner venous transducer. Any help would be greatly appreciated!
|
|
|
Post by dave74 on Jan 9, 2024 15:49:31 GMT
I moved this post to the Fresenius section.
Is the venous alarm the only alarm test that is failing? Did you try a different level detect module? What about the cable that runs from the card cage to the level detect module?
|
|
|
Post by spartaninohio on Jan 9, 2024 16:41:38 GMT
It is the only alarm and I have not tried the level detector module. I will try that.
|
|
|
Post by spartaninohio on Jan 9, 2024 17:06:19 GMT
When I was moving the machine I noticed a small jumper off one of the boards on the floor, no idea how it came off, so I started looking at the venous pressure board and compared it to another "good" board and put it on the missing spot, B6, and it works now. Thanks for your input to get me looking!
|
|
|
Post by dave74 on Jan 9, 2024 17:50:15 GMT
Thanks for posting your fix.
|
|