Post by KimiRaikkonen on Jan 18, 2024 18:47:51 GMT
2008T with 16k hours. I was hoping it was just the filter itself, but I replaced it with a different lot number and ran tests in Dialysis mode and it passed about 8 times with no issue. Do one chemical disinfect and it fails again. The troubleshooting guide seems to only really worry about passing this test in dialysis mode. No obvious leaks, if anyone has any suggestions on where to start it would be greatly appreciated.
|
|
|
Post by aycdit on Jan 18, 2024 19:46:19 GMT
I had this issue once some years back and it was the UF pump.
|
|
|
Post by dave74 on Jan 18, 2024 21:28:58 GMT
Once you bleach the machine is the Diasafe damaged? If you moved that same Diasafe to another machine, would it pass?
|
|
|
Post by gfreely on Jan 19, 2024 17:34:22 GMT
I had this issue once some years back and it was the UF pump. Same here. Slaved one in, did a quick cal and BAM...went through. Believe it pulls the pressure for the test and must have been just over the edge of the limits. Cal'ing it previously didn't fix the issue though.
|
|
phillipashleyludlow
Full Member
01001000 01101001 00100000 01010011 01110100 01110101 01100110 01100110 00101110
Posts: 151
|
Post by phillipashleyludlow on Jan 19, 2024 17:39:56 GMT
*Here Here, I Concur*
UF Pump. alot of my older 1 & 2T's do this from time to time, changed the uf springs, recal'd , passed test, RTS.
|
|
|
Post by gaster75 on Jan 19, 2024 17:41:14 GMT
Make sure you check valve 28 especially wiring. That valve opens to release pressure during diasafe test after bleach.
|
|