Post by ryanp930 on Jan 20, 2024 17:04:18 GMT
Hey Fellas! I had a machine pulled for "TMP not Resetting"
After taking a look at the machine the TMP and Venous was high and wasnt resetting. I changed my Diasafe since it was due anyways. I Rinsed the machine and then put it back in Dialysis mode. That seemed to fix the TMP issue. TMP changed from +200 to +50.
However, I noticed that the Venous Pressure was -100 and was not moving although I tried a reset. Here are a few things I checked and did: 1. I checked the transducer to make sure nothing was in there. It was clear. 2. I swapped out the whole module with one from a good working machine. Still the same problem. 3. I tried a good working Actuator Test Board from a good machine. No luck. 4. I tried the calibration and when I was trying to zero it on the Potentiometer it did not budge. It did change the value of the Actuator which went from 6 to 0 when I adjusted it. Venous number did not change at all.
Not sure if I missed something but if you guys could suggest another thing to try and shed some light I would greatly appreciate it. Thanks in advanced fellas!
|
|
|
Post by dave74 on Jan 20, 2024 17:57:52 GMT
Did you try the cable that goes from the back of the card cage to the level detect module?
|
|
|
Post by ryanp930 on Jan 20, 2024 19:15:34 GMT
Did you try the cable that goes from the back of the card cage to the level detect module? Hey Dave,
Just tried it and its the same issue. I was working on it yesterday when i did my troubleshooting but at the moment the Venous is still -100 and TMP is at +120.
|
|
|
Post by gnurk on Jan 21, 2024 13:23:39 GMT
try your sensor cable then in t andc mode your sensor board
|
|
|
Post by Fresniusguy on Jan 22, 2024 5:06:03 GMT
Last time I had this issue it was my function board and my EPROM could be saved. Even though I didn’t do any damage to it. Once I replaced it all and recalibrated everything in T&C mode.
|
|
|
Post by ryanp930 on Jan 23, 2024 15:08:37 GMT
try your sensor cable then in t andc mode your sensor board Hey guys, it ended up being the sensor cable. Problem was fixed when i swapped it out from another machine. Thanks so much for your help. I appreciate you all!
|
|
|
Post by dave74 on Jan 23, 2024 18:29:59 GMT
Thanks for posting your fix.
|
|