Post by Middle of No Where on Feb 8, 2024 13:52:22 GMT
Had a banner pop up during treatment today during treatment, "Temperature Verification". 2008T BlueStar Machine, treatment time dropped about 5-10 mins after the banner disappeared. Has anyone else seen this? Possible causes? Tech support is useless.
|
|
|
Post by dave74 on Feb 8, 2024 14:35:24 GMT
If you look in the operator's manual, page 276, Fresenius calls this an advisory alarm. The machine will stay in bypass for 8 minutes, and then you will get an alarm of in bypass for 8 minutes, and everything will be OK.
I had this problem long ago when we had temp comp turned on because of variable feed water temp for the acute machines. I never really figured it out except it was related to temp comp. Now all the machines have temp comp on. Fresenius probably calls it an advisory alarm because they can't make it go away either.
|
|