Post by Freseniusguy123 on Feb 8, 2024 16:36:01 GMT
2008T non bibag blue star premium with 30k hours. My pre and post condo are off on debug screen. My conductivity is stable up top but on the bottom it reads 0.468 and the temperature is off also. I've tried swapping the condo cells and the temp sensors. Still with no luck and its preventing the OLC test during treatments.
|
|
stuff
Full Member
1111111111111111111111111111111111111111111111111111111111111111111111111111111111111111111111111111
Posts: 230
|
Post by stuff on Feb 8, 2024 16:42:38 GMT
Your post just states you replaced, have you calibrated Temp/Condo, also to include the pre and post temp sensors?
|
|
|
Post by Freseniusguy123 on Feb 8, 2024 16:44:48 GMT
It won't allow me to keeps giving me an error. Barely getting accurate readings on the external meter. Condo on tcd is 14.0 my meter is reading 0.187. It does the same thing when it tries to calibrate both temperature and conductivity.
|
|
|
Post by Freseniusguy123 on Feb 8, 2024 17:34:56 GMT
Made a rookie mistake and swapped the temp sensors with condo sensors. Once I fixed that. I was able to calibrate the cells & temp control. Rookie dumb mistake. Thanks for the help stuff
|
|
|
Post by dave74 on Feb 8, 2024 18:19:58 GMT
Thanks for posting your fix.
|
|