joey
Full Member
Posts: 227
|
Post by joey on Feb 8, 2024 16:42:09 GMT
Good morning all I’m jumping into a pm and do not remember using a 3 way recirc adapter when they showed us the leak test. Either way does anyone have setups they made for this and send me pics or links to onelove12401@gmail.com Anything will be appreciated
|
|
|
Post by SrCusEngr on Feb 9, 2024 14:53:05 GMT
<joey> You are welcome to correct me if I am mistaken, but I believe the 3-way adapter you are referring to is used only with Solcart holder machines. As a part in the new PM kit, Germany has included a three-way adapter, as well as some tubing, to replace the input segment to the bicarbonate piston pump. There have been some problems with long-term bicarb build-up in those barbs causing problems down the road. Replacing the adapter during a PM will reduce or eliminate those problems.
|
|
|
Post by Admin on Feb 9, 2024 19:10:37 GMT
|
|
|
Post by Chuck Weddle on Feb 9, 2024 20:26:41 GMT
The link comes up as page not found.
|
|
|
Post by Admin on Feb 10, 2024 2:25:32 GMT
|
|
|
Post by dave74 on Feb 10, 2024 3:16:53 GMT
Page no longer exists. I tried on two different browsers.
|
|
|
Post by Admin on Feb 10, 2024 19:49:26 GMT
AH! I see why.
Because of the rules I set, it won’t allow anyone to say C-O-C-K so in the link the **** should be replaced by the word C-O-C-K without the dashes of course.
|
|
|
Post by Admin on Feb 10, 2024 20:40:24 GMT
Just removed the filters. Now it works
|
|
joey
Full Member
Posts: 227
|
Post by joey on Feb 12, 2024 13:14:09 GMT
If you watch the video on the portal it shows a funky setup while doing the leak test that has something plugged into the dialysate sample port, as well as what looks like the dialysate line going to the manometer. The tubing in the video is partially blocked by a clamp so I can not make it out and like I said I definitely would have remembered that in class. Any input would be appreciated.
|
|
|
Post by biomedpete on Feb 12, 2024 14:15:22 GMT
It sounds like you are talking about the DF Block leak test. That would be a 4-way stopcock, with a syringe and pressure meter connected to the sample port.
|
|
joey
Full Member
Posts: 227
|
Post by joey on Feb 12, 2024 16:11:41 GMT
It sounds like you are talking about the DF Block leak test. That would be a 4-way stopcock, with a syringe and pressure meter connected to the sample port. View AttachmentOH OK So it sounds like I can use the 5 way I have for the old ABPM test.
|
|
|
Post by Admin on Feb 12, 2024 16:27:13 GMT
Pretty much yup!
|
|
|
Post by SrCusEngr on Feb 12, 2024 18:02:56 GMT
Or a tubing "T" jig with clamps that you use to use for adjusting RVFPE/RVFPA/RVDA/DDE.
|
|