Post by rawb2023 on Feb 12, 2024 19:01:06 GMT
I just replaced the blood leak sensor on 2008T machine and now I'm getting a Blood Leak 1 fail. I have tried a new sensor board and sensor cable but still getting this fail.
|
|
pacnw
Full Member
Posts: 187
|
Post by pacnw on Feb 12, 2024 19:15:51 GMT
did you calibrate the leak and dimness?? sometimes this helps in making that pass. Also look for leaks as a wet blood leak tube will cause the alarms too.
|
|
|
Post by rawb2023 on Feb 12, 2024 19:36:52 GMT
I calibrated both and it still gives * Fail Blood Leak 1. Ill check tube for leaks and go from there.
|
|
|
Post by dave74 on Feb 12, 2024 20:31:48 GMT
What was the original problem?
|
|
|
Post by rawb2023 on Feb 13, 2024 16:27:58 GMT
Originial problem was that I had a "Blood Leak?" in dialysis mode. I did acide and bleached machine and calibrated blood leak, but the same banner came up. Next I replaced blood leak detector assembly and recalibrated. Now I dont get the banner anymore, but now I get " Blood Leak 1" during alarm and pressure tests. I have no air in system. I put a new sensor board and sensor cable and started machine in T and C mode, but still the same problems.
|
|
|
Post by freseniusguy123 on Feb 13, 2024 17:24:27 GMT
Replacing the sensor board fixed my issue when I had the issue.
|
|
|
Post by dave74 on Feb 13, 2024 18:54:28 GMT
Are you sure your blood leak detector is good and nothing went wrong during the installation? I know I damaged at least one while installing it. It sounds like it calibrated OK. Do LEAK and DIMN on debug screen 4 change during alarms test? I don't have a known working machine to compare.
|
|
|
Post by rawb2023 on Feb 16, 2024 18:37:51 GMT
My new blood leak detectors came in and I put one in machine. All tests passed 3x and no blood leak alarm ever. The blood leak detector I put in previously could have been defective.
|
|
|
Post by dave74 on Feb 16, 2024 21:02:54 GMT
Thanks for posting your fix.
|
|
|
Post by Gac_melle on Apr 6, 2024 14:49:52 GMT
Had a similar issue with Blood leak 1 fail on 2008T Bluestar. My fix was replacing Actuator Test Board.
|
|
|
Post by gnurk on Apr 7, 2024 12:10:45 GMT
fill out a product complaint if you had a oob fail you could get a new blood leak for free
|
|