Post by biomedfossil on Feb 16, 2024 15:39:24 GMT
I work for Fresenius and I have a Aqua B system which is about 4 years old. I have noticed when disinfecting the PsP1 gauge located at the bottom of cabinet 2.. I have noticed it is showing 1bar and looking back at my past disinfection it was 2-2.5bars. It is actually flagging when I complete my TMS pm work order. I just wanted to get a little education on the pressure gauge and what may be causing the pressure decrease. I really don't want to call Fresenius since there is so much red tape on getting any answers. Thanks..
|
|
joey
Full Member
Posts: 227
|
Post by joey on Feb 16, 2024 16:59:48 GMT
does it move? bad gauge? I don't know that system but is it after a pre filter?
|
|
|
Post by biomedfossil on Feb 22, 2024 15:23:06 GMT
Sorry for the late response. The gauge is actually tied into the RO. So there isn't a pre-filter or anything. The gauge is pretty steady so I don't think it's the gauge.
|
|
|
Post by SrCusEngr on Mar 12, 2024 21:16:59 GMT
<biomedfossil> How is it tied to the RO? Product output? A little detail can improve our response. Check the flow diagram, and give us some insight?
|
|
|
Post by robertb on Apr 1, 2024 21:20:31 GMT
it's a sign that membranes are fouling/clogging up You can try to do the citric acid clean before your disinfect It may help ...if not the only way to get the pressure back up is replace membranes
|
|
|
Post by Chuck Weddle on Apr 1, 2024 21:31:26 GMT
Just to clarify, he is talking about a Fresenius Aqua B and not a B. Braun Aqua Boss.
|
|