Post by salloum on Feb 16, 2024 15:40:42 GMT
Dear all , i have a problem with (DF pressure test will be repeated ) , how i can resolve that , how i can decrease the Uf volume ?
Thanks
|
|
|
Post by Chuck Weddle on Feb 16, 2024 16:58:42 GMT
What step is it failing?
|
|
|
Post by SrCusEngr on Feb 16, 2024 17:37:06 GMT
<salloum> Put the machine into TSM (DB switch set to position 2). Select Manual Test and Calibration; now select Low Level Manual Test; now scroll down to test 1.27 DFS Pressure Test of Therapy. When you run this test, it is the exact same test as what is used in therapy, but you can run it over and over without all the preparation steps.
To run the test, press Start. Pay attention to the PDA value. in the lower right you will see LLC and a 5 digit number (12010; 12200; 12250; 12300; ...). These numbers represent the STEPS of the test. The LLS 5 digit number is basically a timer/counter. ALL FAILURES will be indicated by single HIGH PRESSURE Failure message. We need to know what STEP the failure occurred at and what is the PDA value at failure.
Last item: Need Software version number and WTC of machine.
|
|