Post by ariel on Feb 28, 2024 13:14:31 GMT
recently replaced bibag hydraulic assy. and bibag interface board. Now getting bibag venting
|
|
|
Post by middleofnowhere on Feb 28, 2024 14:07:31 GMT
BiBag Pressure Transducer needs calibration or replaced
|
|
|
Post by Biomed111 on Feb 28, 2024 15:36:15 GMT
How often are you seeing this error?
|
|
Icebonez
Junior Member
What is better - to be born good, or to overcome your evil nature through great effort?
Posts: 55
|
Post by Icebonez on Feb 28, 2024 15:37:40 GMT
BiBag Pressure Transducer needs calibration or replaced I agree with Middle on this one. Are you getting the "Bibag: Chamber Venting" or the "Bibag: Vent Too Long" message? I was under the impression that the "venting" banner is just an advisory message, only requiring troubleshooting if the "vent too long" banner appears.
|
|