Older 9.12 machine alarming Blood Pump Stationary during treatment. What is usual cause for this?
|
|
|
Post by qman213 on Feb 28, 2024 18:33:40 GMT
Is the blood pump turning during a mock therapy? Is the one-way bearing in the roller working properly? It should only spin freely one way.
|
|
SrCusEngr
Full Member
Posts: 210
Member is Online
|
Post by SrCusEngr on Feb 29, 2024 17:56:47 GMT
The Dialog+ EVO machines incorporated a new feature - Hall sensors. You will find them by the piston pumps (therefore the ball joint has two magnets), the blood pump roller with black arrows (on the back side you will see two magnets embedded into the plastic frame) and the blood pump motor (note that the blood pump motor connector has two extra wires when compared to the other motors (EP, FPE, FPA). Those two wires are a hall sensor output of the blood pump motor to tell the system of the pump motor is rotating.
In the Rev 1.02 Service Manual, check pages 37/38, the Basic Board. Connector P3 is where the BP Motor (BPA) connector plugs in. Just to the right, P48, is the socket for the Speed/Direction Rotation Arterial Blood Pump DZ/DR BPA. this connector is for those two extra BPA motor wires. Now since this just a Hall sensor output, it is basically a signal switching 0-1-0-1... to indicate speed/rotation (you need to see the starting point to determine direction) which equates to 0 volts/5 volts. So if all the wiring to this point is good and connectors fully seated, and the BP motor is working properly, the next step can be found on page 102. When you look at the left side of the DB diagram, you will find Controller/SUP Sensors (input). Notice that one of them is BPA_IMP. The next page describes it as Blood Pump Arterial Motor Hall Sensor (input). So if the motor is outputting a signal, and all wiring and connectors are good. the problem most likely is the Digital Board.
Be sure to let us know what you find.
|
|
carl
New Member
Posts: 12
|
Post by carl on Feb 29, 2024 23:59:57 GMT
I ran machine in mock therapy for 1.5 hour. No further blood pump issue, but got a temporary communication problem alarm that required turning the machine off and back on again. That returned it to treatment mode. Instructions for use states there is no message received from LLS or sent to LLS for more than 5 seconds. Bad digital board?
|
|
SrCusEngr
Full Member
Posts: 210
Member is Online
|
Post by SrCusEngr on Mar 1, 2024 15:29:04 GMT
<carl> How dusty/dirty is the machine? Not necessarily the DB but I would first check the Diabus cable between the motherboard and the BB. While gold plated pins don't 'corrode' because of the environment (dialysate), microscopic material gets on the pins and causes resistance, thus errors in communications. Reseat connectors and boards to "wipe" the contacts.
|
|
carl
New Member
Posts: 12
|
Post by carl on Mar 19, 2024 1:02:59 GMT
That’s what I did just to be cautious. Problem did not occur again. Thanks.
|
|